Removal of a fractured implant screw

Situation

ree

A 47 year old male patient was referred to the clinic with a fractured implant screw.


ree

A Φ1.2mm Perfect guide connected to a PG driver.

ree

Periapical view. For a perfect guide to be screwed into the screw hole, an implant has to have at least 1mm of the screw threads on top of the fractured screw.

ree

A Φ1.2mm perfect guide was tightened to 15Ncm in the screw hole. A perfect guide positions a drill the dead center of the screw and prevents it from damaging the threads of the implant.

ree

A Φ1.0mm reverse drill is used at 2,000rpm with copious water irrigation. Drilling direction should be in reverse.

ree

Implant motor was turned on reverse.

ree

Using a perfect guide, a reverse drill was placed right on top of the fractured screw. Then approximately 1.0mm was gently drilled vertically down. The use of rubber stoppers is suggested for 1.0mm of depth control.

ree

A Φ1.0mm screw remover matching to the reverse drill is being connected to a shank driver.

ree

The fractured screw was being removed by rotating the screw in counterclockwise direction.

ree

The fractured screw came out of the screw hole.

ree

Drilled site in the center of the fractured screw.

ree

Postoperative view of screw removal.

click on the image to see clinical video.